RE: Dealing with autofluorescence

From: Gerstein, Rachel (Rachel.Gerstein@umassmed.edu)
Date: Fri Jul 27 2001 - 07:17:21 EST


Tony,

Dead and dying cells have higher autofluorescence - try resuspending your
cells in 1 ug/ml propidium iodide (diluted in your staining media; I
recommend RPMI without phenol red + 3% serum for better viability [other
things in RPMI can contribute to autofl when staining with mABs - write me
if interested in the formulation we and many others use]) and gating out PI+
cells in FL3.   Also, larger cells (macs and grans) have higher
autofluorescence than lymphocytes - so even if you have great viability you
may still have the diagonal "tail".  Try using scatter gating to focus on
your cells of interest.

Good luck!
Rachel

=======================================================
Rachel M. Gerstein, Ph.D.
Department of Molecular Genetics and Microbiology
Graduate Program in Immunology/Virology
University of Massachusetts Medical School
55 Lake Avenue North
Worcester, MA 01655-0002
(508) 856-1044
(508) 856-5920 (FAX)


> ----------
> From:		Tony Schountz
> Sent:		Thursday, July 26, 2001 11:13 AM
> To:	Cytometry Mailing List
> Subject:	Dealing with autofluorescence
>
> I'm having a bit of a problem that I'm not sure how to deal with. If I run
> freshly isolated unstained mouse spleen cells through our FACScan, the
> cells autofluoresce equally into FL1 and FL2, to about 10^1.5 or so. In
> other words, there seems to be a perfect diagonal of a few cells that
> autoflouresce (the majority are below 10^1). Gating on any population does
> not affect this pattern. I prep the insturment using Calibrite beads, but
> have not tried adjusting any of the settings to fix this. I've harvested
> spleens in serum-free DMEM (with phenol red), serum-free HBSS (with phenol
> red) or PBS and have washed 2x in 0.5%BSA-PBS, all with similar results.
> Are there any suggestions as to how I can get this under control?
>
> Thanks,
>
> Tony
> --
> Tony Schountz, Ph.D.
> Dept. of Biology
> Mesa State College
>



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:48 EST