Re: Can EYFP be run on FACScan?

From: Joseph Webster (J.Webster@centenary.usyd.edu.AU)
Date: Tue Jul 24 2001 - 00:59:07 EST


Yes, EYFP can definitely be read on a FACScan.
The intensity is less than on the sorter with 514 excitation,
but easily detectable.

For screening clones we run them on the FACScan, only bother
changing wavelengths on the sorter for actually sorting.
If you want to do that, you need 514 filters for FSC & SSC,
also something like 545/35 for FL1; the standard 530/30 will
allow lots of 514 laser light into the fluorescence channel.

Joseph.

At 02:00 24/7/2001, Nancy Harvey wrote:
>Hello,
>I have an investigator who wants to check samples that have enhanced
>YFP.  The ideal excitation is 514.
>Has anyone been able to detect EYFP with the 488 excitation of
>a FACScan?
>
>I could tune the primary laser on the Vantage to 514, but have only 488/10
>filters for FSC and SSC.
>
>Any thoughts and suggestions are appreciated.
>
>Nan Harvey