For bacteria I use ratiometric bead counting and  flow cytometry since the mid
eighties very successfully. For mammalian cells we tend to use the absolute
count facility in the EPICS XL, normally 60 to 100ul (3 or 5* 20ul cycle). The
counting variability actually just done for yeast cells yesterday followed
nicely the statistical counting error (n)^0.5/n in %, e. g 1.46% for 6153 cells
counted (1.29% theoretical for 6000 counts) and 0.74% for 23585 (0.65%
theoretical) for ten sample replicates. Problems come from cells clumping,
sticking to plastic surfaces or sedimenting which also can be a problem for
ratiometric counting. Additional problems come with system dead times if you
use this method.
When you use bead counting ensure you check for static loss and in particular
for bead / cell coincidence of aggregates. It is also not working well if some
of your cells eat your beads. Also consider that the statistical variation of
the ratio is the sum of the two variations.

We actually make our own bead standards and count them out using the EPICS XL
as a very sophisticated 'Coulter counter'


Regards
Gerhard


-----Original Message-----
From:	Karim Vermaelen [SMTP:Karim.Vermaelen@rug.ac.be]
Sent:	Thursday, June 14, 2001 4:32 PM
To:	Cytometry Mailing List
Subject:	absolute cell counts using FACS

x-mac-creator="4D4F5353"
Content-Transfer-Encoding: 7bit

Hi everybody ,
Can anyone recommend a microsphere kit for performing simple cell counts
on numerous samples using the FACS? BD's Truecount beads are only sold
along with a MAb. Our cells do not have to be labelled. Any simpe
alternative? (NB: our good ol' Coulter particle counter is dead)
Tx
Karim

 << File: Card for Karim Vermaelen >>