Re: absolute cell counts using FACS

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Mon Jun 18 2001 - 05:03:23 EST


Hi Karim,

We have been using flow to assess numbers of cells in migration assays
by adding beads to the sample. Basically we are adding a known
amount of unlabelled Calibrite beads (and we know the number per ul by
counting) to a known volume of cells and running through the cytometer.
We then calculate a value for cells per ul from:

[number of cells in area of interest/number of beads collected] x
[bead count per tube/ sample vol in ul]

There are a number of caveats in that your bead counting has to be
accurate, your pipetting needs to be accurate, there needs to be as
constant a flow rate as you can get and the beads and cells need to be
un-clumped. However we have found that this is a good quick method of
seeing certainly relative differences in cell numbers between samples.
I am sure those involved in absolute cell counting will have alternative
approaches!

Good luck!

Derek

On Thu, 14 Jun 2001, Karim Vermaelen wrote:
> Hi everybody ,
> Can anyone recommend a microsphere kit for performing simple cell counts
> on numerous samples using the FACS? BD's Truecount beads are only sold
> along with a MAb. Our cells do not have to be labelled. Any simpe
> alternative? (NB: our good ol' Coulter particle counter is dead)

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk
London, UK			   mobile: 07790 604112

Web Page: http://www.icnet.uk/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:44 EST