Re: CFDA/SE for microbial analysis

From: Richard Haugland (richard.haugland@probes.com)
Date: Fri Jun 08 2001 - 22:49:10 EST


If you are willing to use a different definition of bacterial viability --- membrane
permeability --- then I strongly suggest trying the BacLight Viability kits instead
of CFDA, SE, a reagent whose use is frought with all kinds of potential problems
when used as a "viability" assay, particularly in bacteria. CFDA, SE is a good probe
for looking at cell replication but not particularly good when used for "viability"
due, in part, to the problems that you have observed.

The BacLight kits

http://www.probes.com/servlets/product?region=USA&item=7012

utilize a green-fluorescent SYTO dye for the live population and red-fluorescent
propidium iodide for the dead population.  All one has to do is add the dyes and
observe the fluorescence (or analyze by flow cytometry or in a fluorometer) a few
minutes later. No wash step is required.  The bacteria can be collected on a fliter
before staining, if desired.

Here are some pictures:

http://www.probes.com/servlets/photo?fileid=g000508
http://www.probes.com/servlets/photo?fileid=g000004
http://www.probes.com/servlets/photo?fileid=g000011
http://www.probes.com/servlets/photo?fileid=g000652
http://www.probes.com/servlets/photo?fileid=g000509

and about 31 references:

http://www.probes.com/servlets/bib?item=7007

This assay works in a wide variety of bacteria

http://www.probes.com/handbook/tables/1502.html

You will also not have to remember ANY of the "five things to remember" in the note
below with this type of assay:



> There are five things to remember when using CFDA-se or similar with bacteria:
>
> pH 8 is a very good way to cleave the dye very rapidly, thus you should use
> buffers below pH 7.3
> Ensure that no free esterases are present in your system
> Ensure you switch of the pumps that get the dye out of the bugs before it has a
> chance to bind inside the cell
> Unless you use di-chloro CFDA-SE the intracellular pH can seriously effect your
> fluorescence readout.
>
> Don't say you measure 'viability'!  Esterase activity only indicates metabolic
> activity, not necessarily the ability of the cells to show reproductive growth
> ! You can also use CCFDA-SE to stain any type of vesicle you load with esterase
> which will turn green but is definitively not viable. Assuming from your email
> address that you work for a water company, this is particularly problematic if
> you want to use it in the context of UV treated water where you will create
> DNA, not membrane damage, thus leaving metabolism intact to some extend whilst
> preventing reproductive growth.
>
> regards
> Gerhard
>
> -----Original Message-----
> From:   Daniel Hoefel [SMTP:daniel.hoefel@sawater.sa.gov.au]
> Sent:   Thursday, June 07, 2001 1:58 AM
> To:     Cytometry Mailing List
> Subject:        CFDA/SE for microbial analysis
>
> Dear List,
>
> I would like to pose a question regarding the use of 5(6)-carboxy fluorescein
> diacetate,
> succinimidyl ester (5(6)CFDA/SE).  This question does not relate at all to the
> current
> CFSE discussion on the list.
>
> I was planning to use CFDA/SE to enumerate viable bacteria from water samples by
> flow cytometry, where viability is obviously based on the presence of active
> internal
> esterases that cleave the molecule into its fluorescent carboxy fluorescein
> derivative.
> I have made a stock solution of CFDA/SE (25mM) in high quality, fresh, anhydrous
> DMSO (stock solution stored in a desiccator).   I have been trying to optimise
> the
> staining conditions by testing different buffers and varying pH (buffers
> include PB,
> PBS, MilliQ water and also seeing if the addition of EDTA facilitates
> permeabilisation
> of the cells for better staining).
>
> However, after the addition of CFDA/SE (both 50 and 100uM final concentration)
> to the bacterial suspensions at 35oC, I noticed that after about half an hour
> the
> solutions started to turn a bright yellow/green colour.  I thought that this
> was due
> to the bacteria taking up the molecule and cleaving it internally, and as a
> result
> some of the fluorescent product was leaking out of the cells into the
> supernatant.
> I examined the suspensions by epifluorescent microscopy (blue light excitation)
> and I was nearly blinded by the intensity of the background (ie. the
> supernatant was
> fluorescent, tentatively indicating that the yellow/green colour of the
> suspensions
> was due to the cleaved fluorescent product of CFDA/SE).
>
> Having seen this, I set up a series of negative controls, ie. all the different
> buffers
> without the addition of the bacteria.  I then added CFDA/SE at the same
> concentrations as
> before to the negative controls.  To my surprise, after half an hour at 35oC
> incubation,
> the negative controls started to turn the yellow/green colour.  To further
> investigate
> this, I set up the negative controls once again, this time in a real-time PCR
> machine
> set to detect fluorescein fluorescence.  After the addition of CFDA/SE to the
> negative
> controls, the real-time PCR maching started to detect fluorescence in the
> phosphate
> buffer [pH 8.0] negative controls within 30 seconds.  After 2 mins, all of the
> buffers
> were producing a fluorescent product and within 5 mins the PB solutions had
> saturated
> the fluorescent detector.  The fluorescent intensity and rate of fluorescent
> product
> production was proportional to pH of the buffers where the buffers at a pH of
> 8.0
> produced the fluorescent product faster than buffers at 7.5 and 7.0.
>
> Therefore, it seems clear to me that the addition of CFDA/SE to the aqueous
> buffers is
> causing the molecule to break down into its fluorescent derivative.  The
> manufacturer
> states that if CFDA/SE is stored in the presence of even minimal water, it will
> break
> down over time (therefore CFDA/SE is obviously not stable in an aqueous
> medium). And even
> when stored appropriatley, CFDA/SE is said to only be good for 6 months but
> this isn't
> an issue with my work because I am using new CFDA/SE and storing it in
> anhydrous DMSO.
> Therefore, it appears to me that the molecule is breaking down upon the
> addition of
> it to an aqueous solution such as the buffers.  If this is the case, obviously
> I can't
> use CFDA/SE for bacterial viability studies.
>
> I have read many papers that use CFDA/SE for bacterial viability studies and I
> am
> basically following the same protocol as they report to be successful.  I was
> therefore
> wondering if anyone has seen this occur before or does anyone have any idea as
> to why
> this is happening or what infact is happening.
>
> Have I just got a 'dud' batch of CFDA/SE from the manufacturer?
>
> Any comments or suggestions would be appreciated.
>
> Daniel Hoefel



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:43 EST