Re: Primary Human cells

From: Robin Barclay (robin.barclay@ed.ac.uk)
Date: Thu May 17 2001 - 08:30:49 EST


Obviously the UK and US definitions of Class 1, 2 and 3 are different - and
reversed.
Robin Barclay

----- Original Message -----
From: "Simon Monard" <smonard@trudeauinstitute.org>
To: cyto-inbox
Sent: Thursday, May 17, 2001 2:17 PM
Subject: Re: Primary Human cells


On the contrary!

A Class I biological safety cabinet is a negative pressure ventilated
cabinet usually run with a face velocity of 75fpm or more. The cabinet is
exausted either through a HEPA filter or to the outside. So, a Class one
cabinet would protect the operator but would NOT help keep the sample
aseptic.
A Class II biological safety cabinet on the other hand has HEPA filtered
downward vertical laminar airflow for sample protection and HEPA filtered
exaust for environmental protection.
 I refer you to the CDC Biosafety in Microbiological and Biomedical
Laboratories handbook for such definitioins.




Simon Monard
FACS Lab Manager
Trudeau Institute
Saranac Lake
NY12983

Ph 518 891 3080 X352


>>> "Robin Barclay" <robin.barclay@ed.ac.uk> - 5/16/2001 5:51 AM >>>

Surely Class 2 would be a minimum requirement for operator safety?  Class 1
might keep the sample aseptic, but wouldn't protect the operator.
Robin Barclay

----- Original Message -----
From: "Simon Monard" <smonard@trudeauinstitute.org>
To: cyto-inbox
Sent: Tuesday, May 15, 2001 2:59 PM
Subject: Re: Primary Human cells


>
> I would not  sort primary human or primate material without some sort of
aerosol
> containment, wearing a mask for more than an hour or two is very
uncomfortable. I have
> a custom Class one  type hood on my Vanatge
>
>
>
> Simon Monard
> FACS Lab Manager
> Trudeau Institute
> Saranac Lake
> NY12983
>
> Ph 518 891 3080 X352
>