RE: avidin-FITC intensity increase

From: Robert Archer (robert.archer@probes.com)
Date: Fri Apr 20 2001 - 15:33:41 EST


Fluorophores conjugated to avidin and streptavidin may be quenched
significantly, apparently because the dyes interact with amino acid
residues in the biotin-binding pocket. Exceptions include Cascade Blue–
and phycobiliprotein-labeled avidin and streptavidin; the dyes in these
conjugates are not quenched because they do not interact with the
biotin-binding site. A significant recovery of the avidin or
streptavidin conjugate's fluorescence can be obtained if biotin  is
added as a final incubation step in the staining procedure (see figure).
Fluorescence enhancement of avidin conjugates by biotin has been shown
to occur in <100 milliseconds.1  Biotin apparently blocks the
interaction of the fluorophore with residues in the biotin-binding
pocket that quench the fluorescence, enhancing the fluorescence of the
stained tissue, often multifold.

Figure. Spectra showing the fluorescence of 1) fluorescein-labeled
avidin, 2) fluorescein-labeled avidin after addition of 10 µM biotin and
3) free fluorescein at the same concentration as the fluorescein label
in the avidin conjugate

[Image]

Reference:
1. " Imaging of endosome fusion in BHK fibroblasts based on a novel
fluorimetric avidin-biotin binding assay." Emans N, Biwersi J, Verkman
AS. Biophys J 69, 716-728 (1995)