Re: MEAN CHANNEL VALUE

From: Nancy Perlmutter (ngp@shapirolab.com)
Date: Wed Mar 21 2001 - 10:28:35 EST


Regarding the use of mean channel values on log data:
When data is collected using log amplifiers and displayed on a log scale it
makes sense to use the median rather than the mean channel values as a
method of measuring changes in fluorescence intensity. When data displayed
on a log scale looks like a normal distribution it means that the
underlying linear data is actually skewed toward lower intensity values.
The problem with calculating the mean using log channel values is that it
the mean log channel value is not the same as taking the log of the mean
channel values. The median is a more robust measurement particularly when
working with log channel values. For a much more complete explaination of
the use of mean, CV, median, and robust CV see pages 194-195 of the 3rd
edition of Practical Flow Cytometry by Howard Shapiro. If we are lucky when
Howard returns from his current meeting he will post a much more complete
answer.

Nancy Perlmutter

At 03:47 PM 3/19/01 -0500, Falco, Vincent wrote:

>Group
>
>I have an investigator that wants to use mean channel value on a log scale
>as a measure of changes in fluorescent intensity and thereby a surrogate
>measure for changes in the surface marker antigen. Can I solicit any and all
>thoughts and criticism.
>Thanks in advance.
>Vin Falco

**************************************************
Nancy Perlmutter ngp@shapirolab.com
Laboratory Manager
Howard M. Shapiro, MD, PC

119 Braintree Street, Suite 102
Allston, MA 02134
phone 617-783-8392
fax 617-783-4750
**************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:30 EST