RE: PMT Voltage Responses

From: oiyt (oiyt@tds.net)
Date: Sat Mar 19 2005 - 06:15:44 EST


Hello Flow'ers,
A couple of points on the BD benchtops.
1. Unlike the FACsort & FACScalibur,  the FACScan's filters are fixed and
not easily replaced.
2. Sample voltage is probably one of the more critical readouts because it
tells you if your sheath tank is properly pressurized. A particular voltage
setting is meaningless because they vary from one instrument to the next.
What is important is that a particular instrument have basically the same
voltage readings from one day to the next. The voltage will vary during the
regulation of sample pressure, so it will move around a little, but not
much. It is a good thing to record this voltage in Hi and Lo flow rate,
also in Standby.  When you move the stage under a sample on the FACS
benchtops to the side position (soft standby), your sample voltage should
go to 10.23v very quickly. This tells you that you have enough pressure in
the sheath tank. If you don't get 10.23v very quickly, then you have a
pressure problem somewhere.
3. Actual sample pressures on the FACS benchtops are 4.6 psi in LO and 5.0
psi in HI. (assuming your instrument is properly setup by service person)
The sheath tank sits at 4.5 psi, so Lo flow rate is .1 psi over tank
pressure and HI flow rate is .5 psi over tank pressure.
4. Specification for flow rates are as follows; 12ul (+/-3) per minute in
Lo and 60ul (+/-7) per minute in Hi.
5. Users have no control over the amount of pressure placed on the sample
in the FACS benchtops. As soon as a sample is placed on the SIP (sample
introduction probe), a burst of about 9 psi is utilized to initiate flow.
Sample pressure is then regulated to HI or LO depending on your setting. It
is this bursting phenomenon that requires one to wait for 20 seconds after
placing the sample on the SIP before clicking aquire, thereby allowing the
flow rate to stabilize and become uniform.
Happy flowing,
Tony Leger
Automation Lab Technology
360-983-8690
oiyt@tds.net


-----Original Message-----
From:	Marcus Reckermann [SMTP:recker@ftz-west.uni-kiel.de]
Sent:	Friday, March 16, 2001 1:51 AM
To:	Cytometry Mailing List
Subject:	PMT Voltage Responses


Taking up Bunny?s suggestion, I will share with you the answers I received
to my "PMT voltage" question (in chronological order). The question was
whether the PMT units (mistakenly also threshold) are actually VOLTS or
arbitrary units. Thanks to everybody who responded.

- Threshold is not a voltage setting, it is a channel number
and depends on the sensitivity, alignment, and gain setting of
your particular instrument.
In our FACScan, we had a common measurement with the FSC gain
at 1.75 and threshold 120; after the instrument was serviced
and optics cleaned and realigned, the same result was obtained
with FSC gain of 1.0 instead of 1.75 on the same instrument.
I prefer a description such as: "threshold at 120 in FSC, to
include small lymphoid cells and to ignore most red cells".
A PMT setting of "FL1 640V, Gain 1" is useful, but I also need
to know what optical filters are in that channel.
For instance, in a FACScan the FL1 filter is usually 530/30,
but this may have been changed for this particular experiment.

- Just be aware that voltage and gain settings as well as thresholds are
not
directly transferable between instruments, neither are threshold values.
There
are variations in PMT characteristics and even more important spectral
characteristics of the optical arrangement. In the Elite the threshold is
expressed as a channel number between 0 and 1023, indicating that the
threshold
is dependent on the voltage and gain settings of the PMT

- I thought Threshold was in "Channel Numbers"--at least for the BD
instruments.

- If your data acquisition software writes list mode files that comply
with the FCS specification then you will find a series of keywords and
their values that take the form $PnV and $PnG where n is the parameter
number and V and G stand for PMT voltage and gain, respectively.  With
log amplification, the gain will typically be set to 1 and the voltage
applied to that PMT will determine its sensitivity -- the higher the
voltage the greater the sensitivity.  The second type of description
that you mentioned, i.e., something like FL1 640V, gain 1, is an
appropriate form.  The threshold value is something different and refers
to the signal level in some parameter that must be satisfied in order to
trigger the electronics to capture the signals from that event.  The
threshold value is commonly set on forward scatter which is also most
commonly set to acquire in linear mode thereby allowing various gain
settings other than 1.
 Therefore, to describe the instrument settings, one should specify
the voltages applied to the PMT's ($PnV values), the gains ($PnG
values), log or linear amplification for each parameter and the
parameter used for triggering or thresholding.	 Coulter and BD also
permit one to set a "listgate" or "live gate", respectively, that will
limit the data actually saved to the list mode file.  These gates are
commonly set on the projection of forward vs. side scatter or on
fluorescence from anti-CD45 and side scatter.  If such a collection gate
is used, it too should be specified, at least the combination of
parameters that was used.

- As far as I know, the PMT settings on BD cytometers are reported in
volts,
so you could say "the FL1 PMT was set to 500V" or "the FL1 PMT voltage was
500". Threshold is NOT reported in volts. You may be able to use a formula
to convert the threshold setting into a voltage, but I think the setting
is more universal than voltage for reporting a threshold. BD usually uses
8 volts full-scale, so I would assume that setting the threshold to 1024
would be 8V (i.e., voltage = 8 * 1024 / setting). Since this is not common
knowledge, reporting threshold as a voltage would be meaningless for most.
(Actually, I think on linear scales the 8V is set to channel 1000 and not
1024 because before the FACS Vantage the BD sorters used only 3 digits
[i.e., 000 to 999] to set manual sort gates.)
Also, since the settings on one cytometer don't really mean anything on
another cytometer it doesn't make sense to try to convert settings into
some sort of "real" unit. Simply recording the settings as you see them on
the cytometer is usually the best. "The threshold was set to 20." (Notice
the lack of units.)
One thing that annoys me is the "Sample Pressure" reading on BD benchtop
cytometers (FACScan, etc.). It is displayed as "Volts", because the
cytometer uses an electronic device that changes voltage corresponding to
pressure changes. (The same device can be used in digital tire pressure
gauges.) BD decided not to calibrate these devices, so instead of reporting
an actual pressure they report the voltage of the electronic device. It
makes training difficult when you have to say things like "try to keep the
sample pressure around 6.15 volts".

- Yes, the PMT voltage is volts applied to the PMT.
As you probably know, the sensitivity of the PMT is proportional
to the voltage applied. BD instruments have a maximum PMT voltage of 1000,
I
believe
some other brands go up to 1500 or 2000 volts; most common PMT
tubes have a maximum rating of about 1500 volts.
I don't know the accuracy of the voltage reading, but I don't
think it's important; I use published PMT volts as a starting
guide only, because PMTs vary widely in sensitivity.

- These are actual volt units.	If you set 700volts for a PMT on a Vantage
and
check the cable going to the PMT, it will show	-700 +/- 5V if properly
calibrated.

- it's volts.

- Documenting the conditions under which an experiment is run is a
always a good idea, but it is necessary to understand its use.
Primarily, especially in conjunction with a well understood standards
such a bead, the settings can help someone who knows the flow
instruments at a particular site determine whether problems with
experimental data are due to instrument problems or sample problems.
Secondarily, these settings will allow an instrument user to recreate
the conditions under which previous data were run. Because of
alignment changes between runs, the instrument usually will not give
the same results, but will be a good starting place for setup. And
thirdly, knowledgeable users at other sites may also use these
settings as starting points for their own work, but the variability
from instrument to instrument is too high to do this successfully
without some other portable standard, such as a bead.
The actual instrument values set are "real" electronically, but may
not have much meaning to a researcher. There is too much detail to
include here - I would recommend reading one of the many good books
on flow cytometry that describe the instrumentation.
Briefly, the threshold levels may be a percent of full scale, a
channel number, or turns on a knob. PMT voltages determine the
amplification of the photomultiplier tube that is converting photons
to electrons, which is proportional to the voltage raised to some
power, and varies from PMT to PMT. The gain describes what kind of
electronic amplification is following the PMT, either linear with a
certain gain factor, or logarithmic. Thus, your example of "FL1 640V
gain1" say that measurment channel is in linear gain 1 with its PMT
set to 640V.

- The PMT volt number, is the measurement of the actual voltage applied to
the PMT.  Depending upon the configuration of the PMT, this may be a
positive or a negative voltage. If the voltage is applied to the
photocathode (e.g. BD Benchtops) then it will be negative (with respect to
the anode), repelling the photons converted to electrons, to the anode
(output). If the voltage is applied to the anode, it will be positive (with
respect to the photocathode) attracting the electrons to the output. The
operating range of most flow cytometery PMT's is in the 300 to 1000V range.


___________________________________________________________
Marcus Reckermann
Forschungs- und Technologiezentrum Westkuste der Universitat Kiel (FTZ)
(Research and Technology Centre Westcoast of Kiel University)
Hafentoern
D-25761 Buesum
Tel. +49 (0)4834-604-204 or -261
Fax. +49 (0)4834-604-299
E-Mail: recker@ftz-west.uni-kiel.de
FTZ Web Site: www.uni-kiel.de/ftzwest/
FTZ Flow Cytometry: www.uni-kiel.de/ftzwest/flow/flowhome.htm
________________________________________________
Keep your computer busy with www.setiathome.ssl.berkeley.edu



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:29 EST