RE: Antibody concentration

From: Plett, P Artur (pplett@iupui.edu)
Date: Fri Mar 16 2001 - 12:14:08 EST


Hello flowers, especially those interested in the temperature-antibody
incubation discussion. First of all, thank you for those very educational
essays on this issue.
Now, since the "golden standard" of 1 ug/million cells, which of course we
all know is a titreable industry standard, is expressed in amount of
antibody/cell:
Question 1: how important is the total amount of antibody? Can cells
influence the concentration of antibody in the solution by having it bound
to the cells (if the antibody isn't in a 100-1000 fold excess)? As mentioned
with the parenthetical comment "within reason" this is a frequent situation
with sorting tubes where one is staining up to 10^8 cells and one really
doesn't want to use up half the vial in one shot.

On a more personal matter, in some experiments I did, I expressed the amount
of chemokine receptor staining as 1ug/10^6 cells, which was done in the same
volume (round bottom-tube dry pellet), but with different amounts of
antibody depending on the number of cells that were required for the
experiment.
Question 2: is it appropriate to express it that way (ug/x-cells)?
Question 3: is it valid to compare results if in one experiment 100,000
cells were treated this way (0.1ug/50ul), but in the next experiment 5
million cells (5ug/50ul) were incubated identically? We think that the
amount of antibody (or is it concentration) plays a role. From some in vitro
data we have it seems that the effect we're looking at was there in both
groups, but in vivo data has been very variable (which of course is
characteristic for in vivo data anyway)....

thank you for any input and excuse the "multiparameter" question :)


**********************
Artur Plett, PhD
Post-doctoral fellow
IU School of Medicine Div. of Hematology/Oncology
Indianapolis, IN 46202

-----Original Message-----
From: Donnenberg, Albert [mailto:donnenbergad@MSX.UPMC.EDU]
Sent: Thursday, March 15, 2001 12:22 PM
To: cyto-inbox
Mailing List <cytometry@flowcyt.cyto.purdue.edu> \"Cytometry Mailing
List\"\""
Subject: Antibody concentration



FLOWers-
I would like to emphasize a practical point arising from Simon Hunt's
excellent disquisition on antibody binding (buried somewhere in the middle).
It's the Ab concentration that matters and not the amount of Ab per cell!
The practical ramification is that if you stain in microtiter plates, the
volume of a dry cell pellet (as dry as you can get it by flicking and
blotting) containing 100,000 to several million cells is only about 12
microliters.  Thus you can use about a tenth of the amount of Ab recommended
by manufacturer's protocols that advise resuspension in 100 microliters.
Practically we use from 1 to 3 microliters of antibody per test, depending
on how a specific antibody titrates and on how much the antibodies dilute
each other in multi-color combinations.

It also follows that (within reason) a very large number of cells can be
stained in a very small volume with little reagent.  This is helpful for
sorting.

Albert Donnenberg