Re: YFP

From: Andy Johnson (acj@mrc-lmb.cam.ac.uk)
Date: Thu Mar 15 2001 - 11:14:24 EST


Hi Jacqueline,

I suppose it is down to which lasers you have available.  On a standard
BD vantage the UV and 488 lines are most accessible so the obvious
pairing is GFP with YFP or DsRed (RFP).  If you have a tunable argon
laser (as on a MoFlo) then the 457 line is fine to excite CFP and YFP
pairs, although this will need compensating and fine tuning which
filters you use. The BFP doesn't seem to be the prefered choice but it
does allow for the largest spectral separation - BFP with GFP or YFP.

Also don't rely too much on the "published"  Excitation/Emission
profiles of the GFP family, as they are most often different and require
alittle bit of flexibility.

Andy




This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:57:28 EST