Re: Need a simple procedure for PI/Cell surface Staining

From: Corrado.Cilio@cmb.umu.se
Date: Fri Mar 09 2001 - 22:49:50 EST


Bill,

I do not think you need to use ETOH if you have fixed your cells with PFA
for membrane staining and PI cell cycle. After surface staining and fixation
resuspend the cells in 0.1% Na citrate, 0.1% Triton X-100, 50 ug/ml PI and
eventually RNAse 50 ug/ml. Leave cells ON and then analyse by FACS. That
will make the job. Remember that with PFA fixation PI profiles do not look
as good as if you use ETOH to fix. On the other hand surface staining is in
most cases affected by ETOH fixation. You have to find the compromise which
best fit your purpose.

Hope it helps

Corrado