Re: MLUV & Indo-1

From: Howard Shapiro (hms@shapirolab.com)
Date: Thu Jan 25 2001 - 20:33:37 EST


Andrew Beernink asked:

>If you're running X mW of MLUV (light regulated), will a shift in power
>among the UV wavelengths change the ratio of Indo (high:low), or
>proportionally raise/drop the intensity of both Indo wavelengths I'm
>collecting (405 bp & 485 bp)?  In other words:  in light regulation mode,
>based on changes in current, one might get an increase in wavelength A and a
>decrease in wavelength B.  The question posed is, how will this change
>fluorescence?

There are substantial differences between the excitation efficiencies of
indo-1 with and without bound calcium at 351 and 363 nm (the ratio of
excitation at 351 to excitation at 363 changes more for the
calcium-saturated than for the calcium-free compound; see the Molecular
Probes handbook), so changes in the ratio of these two lines while
multiline UV power output remained constant would be expected to cause
changes in the 405/485 emission ratio in the absence of changes in
intracellular calcium.  It would be preferable to use single-line UV.

-Howard