Re: No flow on cytometer

From: Mark Kukuruga (kukuru@med.umich.edu)
Date: Fri Jan 19 2001 - 10:31:49 EST


Tony,
A question . . . do you see no flow at all?  If this is true, then check the lines
going through the pinch plates on the back of the control valve.  I used to find that
after periods of disuse these tubes would be stuck together (crimped?), even after
there was no pressure from the plate (hope this is clear).
Alternatively, check that there is no blockage to the waste tank, and the waste tank
is vented.
If you're getting flow to the waste tank, but cannot get sample to flow, check the
sample tube seal.
Finally, you can also apply back pressure to the sipper . . . remove the drip-back
vacuum sheath, place a bit of small tubing on the tip of a syringe, and attach that
to the end of the sipper.  Be careful not to torque the sipper - - this goes directly
into the flow cell.  Disrupting its alignment into the flow cell will dramatically
effect your machine's alignment.
MAK.

--
Mark A. KuKuruga, Managing Director
University of Michigan Flow Core
7416 CCGC 0946
(734) 647-3216, fax (734) 936-7376
kukuru@umich.edu


>>> Tony Schountz <tschount@mesastate.edu> 01/17/01 09:32PM >>>

Our BD FACScan seems to have flow problems. It hasn't been used for a
couple of months, so I think it's just clogged lines. But since we don't
have a service contract I was hoping the group might have some
suggestions as to get the lines unclogged. In the event this isn't
the problem, what else should I check out?

Thanks,

Tony

--