RE: cyan fluorescent protein...

From: Howard Shapiro (hms@shapirolab.com)
Date: Thu Dec 14 2000 - 19:44:50 EST


Kevin Homes wrote, in response to Andrew Beernink's query about excitation
for cyan fluorescent protein -

>We've used the 413 line from our krypton in recent studies (Science Jun 30
>2000: 2354-2357 and (www.stke.org/cgi/content/full/OC_sigtrans;2000/38/pl1);
>Roger Tsien suggested that the HeCad 440 line is optimal, but this was for
>microscopy.  I'd like to see if this is true for flow as well, but I don't
>have a HeCad to try at the moment.

The excitation spectrum is what it is; if 440 nm is optimal for microscopy,
it should be for flow as well.  Note that most of the He-Cd lasers used in
flow cytometers emit at 325 nm and do *not* have interchangeable mirrors,
so you'd need a new 440 (closer to 442, actually) nm He-Cd laser.  But you
can get 100 mW of 442 nm excitation out of a He-Cd laser which is
air-cooled and runs on 110 VAC, although it will cost you $20K or so.

-Howard