RE: cyan fluorescent protein...

From: Pizzo,Eugene (Pizzo@nso1.uchc.edu)
Date: Mon Dec 11 2000 - 20:44:03 EST


Hi Andrew,

I just tried this for the first time last week. And beginners luck I guess,
worked great.
I was doing CFP/YFP and for the YFP excited with 488 and picked up with a
550/30 BP
After the standard 560 SP dichroic into FL1. For CFP I excited with a 406nm
Krypton line
And picked up with a 485/22 BP past a 505SP into the FL4 channel on the
Vantage.
Separation of both signals from the double positive was excellent with
little or no 488
Entering FL4 from the first laser - so I was thankful I didn't have to shift
the first laser into
The 512nm line and lose my scatter or use only one laser at 458 and lose
scatter and have compensation issues.

Of course I thank Santa for the Krypton laser.


Gene/ http://flowcytometry.uchc.edu <http://flowcytometry.uchc.edu>


	-----Original Message-----
	From:	Andrew Beernink [SMTP:ABeernink@novasite.com]
	Sent:	Friday, December 08, 2000 12:04 PM
	To:	Cytometry Mailing List
	Subject:	cyan fluorescent protein...


	Hi-

	What are people using to excite Cyan Fluorescent Protein?  We have a
457
	line, but unfortunately it excites other dyes we're using.

	Anybody using a 440 HeCd, or even better, a 413 Krypton line?

	Thanks.

	Andrew Beernink
	Novasite Pharmaceuticals