Re: Venous versus cardiac blood collection in mice

From: Kate Dugan (K_Dugan@hendrix.jci.tju.edu)
Date: Wed Dec 06 2000 - 13:32:20 EST


Richard

Have you tried just adding more heparin?  It is recommeneded to use 20
units-50 units per mL of blood.  Also, it is extremely important,
especially with larger quanities of blood, to ensure the blood is inverted
several times to make sure it is evenly mixed with the anticoagulant.  I
collect blood routinely via the orbital sinus and add 20 units of heparin
to 300 microliters of blood.  I only used to add 10 units, but I was having
a clotting problem.  I know this sounds too simple, but maybe it will solve
your clotting problem.

Kate


At 04:36 PM 12/4/2000 -0500, Bill Telford wrote:
>
>Hello all...
>
>My former postdoctoral mentor has this question regarding blood collection
>from mice for flow cytometry.  Any ideas?
>
>Thanks!
>
>Bill Telford
>
>_______________________
>Richard A. Miller (U. Michigan) asks...
>
>Got a question for you about mouse blood and flow cytometry.
>We've been doing T cell subset on mouse tail blood for a while
>now. The protocol involves heating the mouse under a light bulb;
>then making a nick in the tail; then collecting 400 ul of blood into a
>tube with a 10 ul drop of heparin in it; then washing with 600 ul of
>PBS; then resuspending to a volume of 400 ul using BSS-BSA;
>then aliquoting and adding antibodies. The flow rate has been 100 -
>200 events/second using our FacsCalibur; we typically count
>10000 events over 50 - 100 seconds.
>We've now been doing the same thing but using cardiac blood
>instead, and we get a much slower flow rate. The mice are
>anesthetized (avertin); heart puncture using a 26 gauge needle
>provides 600 - 1000 ul of cardiac blood. The needle is removed and
>the blood is expressed into a tube containing 10 ul of heparin. We
>then add 1.5 volumes of PBS; centrifuge; discard the supernatant;
>add BSS to the original blood volume, and aliquot. But with this
>method we get flow of only about 50 events/second, and
>sometimes as low as 20 events/second.
>I suspect a clotting problem; perhaps the damage to the heart
>muscle, or perhaps by damage to the cells as they come into the
>needle and syringe. The techs that do these bleedings tell me that
>they do occasionally see evidence of slight clotting in the cardiac
>samples, and I'm therefore suspicious that cells are being
>sequestered or damaged during the blood collection process.
>Do any of you have an idea of what's really going on here, or how to
>solve it?
>
>Rich
>_______________________
>
>
>
***********************************************
Kate Dugan  M.S.
Thomas Jefferson University
BLSB 708
233 South 10th Street
Philadelphia, PA 19107
Lab:  215-503-4545
Fax:  215-923-4153
Email:  K_Dugan@lac.jci.tju.edu
***********************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:56:15 EST