I am gratified by the significant interest in multi-color work!  I
hope everyone who is interested does take Nicole's encouragement to
try it!

mr


(PS, regarding the software compensation issue.  Both Rafael and Ray
are incorrect in asserting that I think there's only one package that
does post-hoc compensation.  There are quite a few, dating all the
way back to Bob Murphy's CALC4 program in the early 1980's on the
PDP-11 and VAX systems.  In fact, compensation is not unique to Flow
Cytometry; scientists have been doing spillover correction
(compensation) for liquid scintillation analysis since before I was
born ("1980").

I would urge Ray and Rafael to re-read what I had written:  "there's
really only one package that can handle the complex needs of
multicolor experiments, from off-line compensation of all the
different colors to keeping track of the dozens (if not hundreds) of
distinct gates applied to a single sample."

While many packages can handle compensation, only one can really deal
with the combination of (1) more than 8 parameters; (2) keeping
data-base like track of dozens of gates; (3) automatically applying
compensation matrices to the correct samples; (4) storing the many
different compensation matrices required in a single experiment with
gates, statistics, and analyses; and (5) presenting all of this
information in a manner that can be managed by humans when dealing
with these extremely complex analyses.  FlowJo was originally written
and developed at Stanford to deal precisely with the new hurdles &
complexities of 5+ color immunophenotyping experiments; after it was
commercialized by Tree Star, it has continued to evolve so as to
serve a wide segment of the flow community.

Finally, Rafael asks if off-line compensation is necessary for doing
multi-color work.  For most 3-color applications, and some 4-color,
the answer is "no".  For more than 4 colors, the answer is almost
always yes--and this is because the current generation of hardware is
not sufficiently adept at allowing correct compensation,
easily-defined, and easily-changed on a sample-by-sample basis.)