We occasionally see really weird findings like this, although we
haven't seen CD8 and CD19 coexpressed. What we do in a situation like
this is re analyze the specimen, or in the case of peripheral blood
get another specimen. We would look at the antibody in question in
FITC, PE, and a third color if possible and use 2 different clones (
for example 2 different sources). We would also pair it with
something else- like CD20 if CD19 was used last time. We would also
make sure the blood was washed several times before staining as
artifactual staining can be observed in unwashed blood ( Commun.
Clin. Cytomet. 18:140-146, 1994). If it is still positive under all
of these conditions then we would consider it real. Until then- it
could  be artifact.

	Maryalice

>Has anyone seen a case that looks like classic CLL (dim kappa+, CD5+, dim
>CD20+, CD23+, CD43+, CD10neg, CD3-,CD2-, CD7-, CD4-) BUT with CD8positivity
>that coexpresses CD19?  What do you think?
>thanks, bessie
>--