Re: FACSCaliburs and MacLogin

From: Matthias Haury (mhaury@igc.gulbenkian.pt)
Date: Thu Aug 31 2000 - 04:30:56 EST


Marcia,

Well, what is true is that the FACSCalibur does have difficulties to connect
to a Mac when the Mac has been restarted while the Calibur is already
running for a while... (the time is not really consistent).

This seems to be because the BDIO board on the Calibur must run some kind of
initialisation routine during which it accepts connections attemps from the
Mac, after that - doesnīt work anymore. Only BD knows the details why..

But this is completely independent of MacLogin !!!! Happens with every Mac !
And itīs a BD Problem that they should fix ! As far as I know there is no
way around than restarting the Calibur shortly before the Mac...

Cheers, Matthias
_____________________________________
Prof. Dr. Matthias Haury
Instituto Gulbenkian de Ciencia
Rua Da Quinta Grande N° 6 - Apt.14
P-2781-901 Oeiras Codex
Portugal
Email: mhaury@igc.gulbenkian.pt
Tel: + 351 21 440 79 85
Fax: + 351 21 440 79 70
_____________________________________





> From: "M Wade" <mwade@paramount.bsd.uchicago.edu>
> Date: Tue, 29 Aug 2000 11:15:57 -0500
> To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
> Subject: FACSCaliburs and MacLogin
>
>
> Hello Flowers,
>
> We use MacLogin on our FACScans, FACStar plus, FACSort, and BD LSR and it
> works fine.  Our users are able to login and then restart the computer to
> log off without having to ever turn the cytometer off. On the other hand,
> the FACSCalibur we have will not connect to the cytometer after being
> restarted unless the Calibur is also turned off and then back on.  BD says
> that this is a normal problem and has to do with the incompatibility of
> MacLogin w/ the Calibur.  Has anyone encountered this problem and found a
> way to circumvent it?  If so, please let me know.
>
> Marcia Wade
> Flow Cytometry Technician
>
> University of Chicago
> Immunology Applications Core Facility
> 5841 S. Maryland Ave.
> Chicago, IL 60637
> mwade@paramount.bsd.uchicago.edu
>



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:56:03 EST