Hoechst/BrdUrd/PI method for cell proliferation

I always use

100 mM Tris, pH 7.4, 154 mM NaCl, 1 mM CaCl2, 0.5 mM MgCl2, 0.1% (vol/vol)
Nonidet-P40, 0.2% (weight/vol) BSA, 1.2 µg/ml Hoechst 33258. Prepare weekly
from a 10 x concentrated stock solution in deionised water.

I then add  PI solution at 200 µg/ml in distilled water to give a final
concentration of 2µg/ml.

The permeabilisation procedure is carried out on ice and the cells are also
stored on ice before analysis.

Rabinovitch, P.S., Kubbies, M., Chen, Y.C., Schindler, D., and Hoehn, H.,
(1988). Exp. Cell Res., 74, 309.
Ormerod, M.G., and Kubbies, M. (1992). Cytometry, 13, 678.

The method is discussed fully in M.G. Ormerod & M. Poot. in 'Flow
Cytometry: A Practical Approach'. 3rd Edition. Edited by M.G. Ormerod. IRL
Practical Approach series. Oxford University Press, 2000.

Michael Ormerod
34 Wray Park Road
Reigate RH2 ODE
Telephone: +44 (0)1737 241726
FAX: +44 (0)1737 226736
Mobile telephone: 07802 293242
Web site: http://ourworld.compuserve.com/homepages/Michael_Ormerod

I see no reason why themethod should not work after fixation in ethanol,
but I have never tried this. The standard mwthod os so straight forward I
have never seen a good reason for changing it.

Message text written by "Lindemann, Carsten"
>
I have two questions about the protocol for the BrdU/Hoechst Quenching
assay.
Is it sufficient to permeabilize cells by using NP40 (0,5%) buffer in order
to stain nuclear DNA?
Is fixation with ethanol recommended or does Eth disturb Quenching of
Hoechst dye?

Thanks for any comments

Carsten Lindemann