Re: Apoptosis assays

From: Derek Davies (daviesd2@icrf.icnet.uk)
Date: Tue Jul 25 2000 - 02:54:49 EST


On Sun, 23 Jul 2000, Len Brown wrote:
> What is the consensus as to "wash" or "not-to-wash" human blood cells after
> Annexin-V FITC staining. Most protocols do not wash. If you wash in what
> buffer, Annexin Ca++ binding buffer, PBS etc.? This subject has been touched
> on briefly in the past but not resolved.

Hi Len,

I think that you are right in that most kits that use Annexin do not
generally suggest washing although some do dilute with the annexin
binding buffer before running on the cytometer. It is a bit of a balance
between ensuring that there isnt any 'non-specific' fluorescence due to
the annexin-fluorochrome still being in the tube even if unbound and
losing cells - possibly preferentially losing apoptotic cells, after
washing. In general, with the samples I use, I am more concerned with
cell loss so dont wash and hope that the amount of background will be
the same with all samples. I would be interested to hear how others do
it though.


> Secondly, if staining for apoptotic cells with PI after ethanol fixation and
> subsequent analysis for a sub-G1 peak, do investigators apply the same
> strict criteria as to CVs of G0/G1 peaks as if they were solely performing
> cell cycle analysis, i.e. <3% for a diploid human lymphocyte population?

To see a sub-G1 peak there has to have been DNA damage (by whatever
route) and loss so I would have thought it unreasonable to expect the
same criteria to be applied to the assessment of the G1 peak - even
within the populations that show 'normal' DNA content there may have
been DNA loss to a small degree as well as conformational changes which
will affect the CV of the peak. In general I tend to use the sub-G1
method as a more qualitative than quantitative method, certainly I
wouldnt suggest that ALL the apoptotic cells would appear there, its
good to give you an idea that something is happening but than I would
use other methods to get better quantitation.

Best wishes,
Derek

************************************************************************
Derek Davies                       Voice: (44) 020 7269 3394
FACS Laboratory,                   FAX: (44) 020 7269 3100
Imperial Cancer Research Fund,     e_mail: derek.davies@icrf.icnet.uk
London, UK			   mobile: 07790 604112

Web Page: http://www.icnet.uk/axp/facs/davies/index.html

In tenebris lux
*************************************************************************



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:57 EST