From: Candace Enockson (enockson@musc.edu)
Date: Tue Jun 27 2000 - 16:15:54 EST


After 18 months of use, I had trouble a while back. The BD engineer had to
come out a couple of times, tried adjustments to the laser alignment but
that didn't do it. Changed pressure inducer.  A month later it went out
again, this time we fixed it over the phone by changing the sheath filter.
They told me time delay is usually a pressure problem.

Good luck,
Candace Enockson
Medical Unversity of South Carolina

--On Mon, Jun 26, 2000 5:07 PM -0400 "Warren D. Shlomchik"
<wds8@email.med.yale.edu> wrote:

>
> Dear Flowers (in particular FACS Calibur users),
>
> We have a new FACS Calibur and have had persistent problems with
> reproducibly being able to set time delay calibration.  In
> particular, we get the error message, "signal out of range...unable
> to set time delay."  BD has replaced an EPROM and a board, but the
> problem intermittently persists.  In speaking to users of 3 other
> Caliburs I discovered that they too had seen this problem.  I'd
> appreciate any comments from other two laser users on the Calibur and
> will summarize the responses to the list.  Many thanks.
>
>