>From what I can gather it sounds like you are using a 60Hz AC light source
and doing brightfield imaging.  You will have to find yourself a DC lamp
which won't flicker at high frequencies.  If you are doing fluorescence
work then I would suggest getting a new lamp since they tend to fluctuate
in intensity as they get older.

Other things to consider might be the trigger going to the camera or the
speed with which the camera can be triggered may be less than your IPLab is
sending.  I don't really have enough information here to help you (eg: what
kind of camera, type of controlling card etc...)

My two cents worth.

At 04:51 PM 6/6/2000 +0000, Maris Handley wrote:
>
>Hello Everyone,
>
>We are doing time lapse experiments with a Nikon microscope (using IPLab
software).
>When we play the movie back we are seeing large changes in light intensity
(almost
>random...not getting brighter continuously or dimmer continuously).  Our
exposure time
>isn't being changed.  I believe this is a problem with the light source
stability,
>but is there something else that might be causing this?  Has anyone else
seen this,
>and what did you do to solve the problem?
>
>Thank you for any help,
>Maris
>
>
>Maris Handley
>Dana-Farber Cancer Institute
>Boston, MA 02115
>(617)632-3179
>
>
- Derek Schulze