Re: cytotoxicity testing

From: Mark A. KuKuruga (kukuru@umich.edu)
Date: Fri May 19 2000 - 07:43:57 EST


M,
You really shouldn't have any problem co-staining with a PE moab . . . you may
have to play with concentrations of the viability/cytotoxicity indicators, to
better accommodate the signal crossovers.  Depending on the machine (whether or
not you can change filters), using a narrower bandpass for the PE (I sometimes
use a 575/14) can reduce compensation problems.
If you can go APC, much better.
MAK.