Joseph, Your problem may be that the Transverse mode (TEM) of your I70 has degraded from the 00 to 01 or higher. The transverse mode number represents the number of nulls that appear in the cross section of the beam. As the number increases, the beam diameter as well as the divergence increase. This may result in a doughnut shaped beam. If this happens, your cell now passes effectively through 2 or more beams, which will generate a broader pulse (with peaks and valleys) from your PMT, hence the wider CV's. The most common cause of this problem is the tube sagging over time, although there may be other causes. If your I70 has an aperture try decreasing the setting. Your power output will probably decrease, but your beam quality that intercepts the cell may improve. Ask your service engineer to compare the beam quality of your I90 to your I70. There are a few companies that sell laser beam characterization test equipment that not only measures power, but also measures beam shape as well as intensity profile. Hope this helps. Best Regards, Ray Lannigan Tritech Field Engineering Inc. 2014 Renard Ct. Suite I Annapolis, MD 21401 800-886-7004 410-266-1522 www.tritechinc.com P.S. We are thinking of providing laser beam characterization services for our customers and would like to get feedback from flowers that would be interested ( or not) in that service. Any responses are greatly appreciated. -----Original Message----- From: Joseph Webster <J.Webster@centenary.usyd.edu.AU> To: cyto-inbox Date: Monday, May 08, 2000 5:09 PM Subject: Another laser question... > >Hi All, >Has anyone compared the Coherent I-70c with I-90/5UV for >visible light performance, particularly stability? > >I was running the I-90 as primary laser, 488/200mW, last week >swapped the I-70 into that role. > >I'm not happy with the result.... > >I don't have equipment to measure output power jitter, >but the CV of bead peaks is around 2-3 times more with the I-70 >compared to the same measurement done with the I-90; the FSC >is now lots murkier than before. > >I would expect similar results from the two lasers running in >exactly the same conditions, but the Coherent engineer is not >convinced there is anything wrong. > >Any experience from the real world would be appreciated. > >Thanks, Joseph. >-- >Joseph Webster, Flow Cytometry Facility >Centenary Institute, Sydney AUSTRALIA. >
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:48 EST