I usually use flow-check or flow-set beads from Beckman-Coulter ( they are
at 1 X 10.6/mL). I put 10 uL in 1mL of cell suspension, I stop the analysis
on  a 1000 beads ( equivalent to 100 uL of cell suspension),  I just have to
multiply by 10 the number of cells ( in a specific gate) and I have the cell
concentration per mL.

You just have to be carefull for cell aggregats, that is the same problem
for any counting techniques
Put aside  a vial of beads for that specific purpose, vortex it well before
using and pay attention to tiny air bubbles when pipetting ( specially
following the vortexing ).


		Lucie Bourget

		Flowcytometry
		Biotechnology Research Institute
		National Research Council Canada
		Montréal, QC, H4P 2R2
		Tel: office 514-496-6274 lab 514-496-6330
		Fax: 514-496-5143




		-----Original Message-----
		From:	Arnold Pizzey [mailto:a.pizzey@ucl.ac.uk]
		Sent:	Thursday, May 11, 2000 4:31 AM
		To:	Cytometry Mailing List
		Subject:	True count substitute


		At 09:21 09/05/00 -0400, you wrote:
		>
		>
		>
		>Dear Flowers,
		>
		>BD does not make the TruCount tubes available without
buying it with their
		kits
		>(that we do not need). Does anyone know a similar product
for cell
		counting on a
		>flow cytometer?
		>
		>Thanks,
		>
		>Akos Szilvasi
		>Biogen, Cambridge, MA
		>
		>
		>
		>
		>
		Greetings Akos,

		I have been using the following method to obtain absolute
counts by flow;

		1) centrifuge 5ml of Beckman-Coulter Flow-check
Fluorospheres 400g 10mins.

		2) Discard supernatent and resuspend in 5ml fetal bovine
serum

		3) centrifuge as above and resuspend in 5ml fetal bovine
serum

		4) Count beads by haemocytometer (the hard part - I count
around 1000 beads)

		5) Add beads to sample and analyse by flow.


		regards,

		Arnold


_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/
			Arnold Richard Pizzey
			Department of Haematology
			Royal Free and University College London Medical
School
			98 Chenies Mews
			London WC1E 6HX
			U.K

			voice:	+44 020-7679-6234
			Fax:	+44 020-7679-6222
			email:	a.pizzey@ucl.ac.uk

_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/_/



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:48 EST