From: Marty Bigos (bigos@stanford.edu)
Date: Wed May 10 2000 - 14:33:14 EST


Hi-

We have used an I-90 for simultaneous UV and 488 nm on a FACS Vantage
and found it to be problematic. First, there is no independent
adjustment of the power output of each line, so in order to get good
UV the laser needs to be run at a high current resulting in high 488
nm output. Secondly, the tube life seemed to be much shorter than an
I-90 in single line mode. We also seemed to have more problems with
beam shape and noise, especially as the tube aged.

The Coherent Enterprise laser is designed to simultaneous 488 nm and
UV; we don't have any experience with that model.

If you can afford it and have the bench space, in our opinion having
separate lasers would be optimal.

Marty Bigos
Stanford Shared FACS Facility

>Dear fellow Flowers,
>
>we have ordered a MoFlo Sorter and, concerning simultaneous
>excitation in UV and at
>488nm, whether one large 5 W Argon ion laser giving both lines or an
>additional smaller
>laser for the 488nm is better or is there any even better third possibility??
>
>Thanks a lot in advance
>
>Renate and Isabel
>PEI flowers