Re: apoptosis and monocytes

From: Carol.W.Johnson@ap.pnu.com
Date: Tue Apr 25 2000 - 09:53:29 EST


Has anyone actually looked at the cells with a fluorescent microscope?

We found a similar signal for rat bone marrow cells, but instead of
autofluorescence, the cells had speckled surface staining
with Annexin V.

I agree with Dr Darzynkewicz that microscopy could be of value when an unusual
Flow Cytometry signal is present.

Carol W Johnson DVM PhD
Pharmacia Corp.








DARZYNKIEWICZ ZBIGNIEW <DARZYNK@nymc.edu> on 04/22/2000 11:17:20 AM

To: cyto-inbox
cc:    (bcc: Carol W Johnson/USKZO/PNU)
Subject:  apoptosis and monocytes







Maciej Simm wrote:
>
> One of the people I'm working with right now is studying apoptosis in
> periteneal macrophages using a kit with annexin V monoclonal FITC ab
> and propidium iodine.
>
> The cells are autofluorescent. The "unstained" tube has signal in up
> to 2nd log.
>



Monocytes/macrophages may often be positive (by flow cytometry) for several
apoptosis markers for the reason that they may have ingested apoptotic
bodies detached from genuine apoptotic cells. This is particularly evident
e.g. during aggressive chemotherapy of leukemias when many leukemic and
perhaps normal lymphocytes die by apoptosis. During ingestion of apoptotic
bodies most likely the plasma membrane of the latter fuses with the
macrophages' plasma membrane which leads to exposure of phosphatidylserine
on the surface and annexin V positivity of macrophages. The ingested
apoptotic bodies have fragments of chromatin and thus multiplicity of DNA
strand breaks, which makes them also positive in the TUNEL assay (e.g.see
Bedner et al., Cytometry, 35: 181-195,1999). I am guessing that such
monocytes may also be positive for for other markers e.g. these that detect
activated caspases, cleavage of PARP, etc. The only way to identify such
"false positive" cells is microscopy or Laser Scanning Cytometry, which
allows one to relocate them and examine whether they are loaded with
apoptotic bodies.
Zbigniew Darzynkiewicz