Fwd: Cyclin D1

From: Maryalice Stetler-Stevenson (stetler@box-s.nih.gov)
Date: Fri Apr 21 2000 - 09:45:57 EST


We worked on this and found the detection of cyclin D1 to be highly
technique dependent. You can not viably freeze the cells, exact
procedures must be followed and someone having a bad day- a tiny bit
off in timing, etc- can't reproduce this test. Matter of fact, only
one person in the lab could reproduce this test. Surprisingly, we
found that DNA content analysis was as good at detecting mantle cell
lymphoma and much more dependable a method. Since the primary
differential diagnosis is between CLL and mantle cell and since CLL
doesn't cycle and mantle cell does, you can make a diagnosis. I have
talked to many who have tried and come to the same conclusion- don't
try to use cyclin D1 in the clinical lab. It is a waste of your time.


	Maryalice


>From: "Dorn-Beineke, Alexandra"
><alexandra.dorn-beineke@ikc.ma.uni-heidelberg.de>
>To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
>Subject: Cyclin D1
>Date: Thu, 20 Apr 2000 11:56:34 +0200
>X-PMFLAGS: 34078848 0 1 9869.CNM
>
>
>Dear Collegues,
>we are looking for a simple method to stain Cyclin-D1 for identification of
>mantle cell lymphoma (translocation t11;14) by flow cytometry, ideally in
>whole blood.  Current protocols for cytoplasmic detection seem to fail and
>also those published are difficult to reproduce and need significant amounts
>of blood.  As we get less and less volume of blood this becomes a problem.
>Thanks for your support.
>
>Alexandra Dorn-Beineke
>Thomas Nebe
>
>Dr. med. Alexandra Dorn-Beineke
>FÄ für Laboratoriumsmedizin
>Universitätsklinikum Mannheim
>Abt. Hämatologie, Immunologie, Allergologie
>Theodor-Kutzer-Ufer 1-3
>D-68167 Mannheim
>Tel:  +49 621 383-3561 o. 3486
>FAX: +49 621 383-3819
>e-mail: alexandra.dorn-beineke@ikc.ma.uni.heidelberg.de
>http://www.ma.uni-heidelberg.de/inst/ikc/

Maryalice Stetler-Stevenson
Director Flow Cytometry Unit
Laboratory of Pathology, NCI, NIH



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:45 EST