Re: Reference- TRYPANE BLUE EXCLUSION

From: Howard Shapiro (hms@shapirolab.com)
Date: Wed Mar 29 2000 - 21:21:42 EST


Peter Kierulf wrote:


>We as many others have been using trypane blue staining to account for
>number og dead cells.
>We have tried in vain to find a basic reference that gives information as
>to why the cells become trypane blue positive when they are dead.
>Can anyone help out??

Trypan blue staining works pretty much the same way propidium staining for
dead cells works; it depends on loss of membrane integrity in dead
cells.  Acid dyes, such as trypan blue, and dyes with double positive
charges, such as propidium, are normally excluded by intact cell
membranes.  When there is sufficient damage to the membrane, these dyes can
enter the cell.  Acid dyes such as trypan blue bind to intracellular
protein; dyes such as propidium bind to nucleic acids.

In general, cells in a culture which are permeable to trypan blue or
propidium are dead; however, cells which are reproductively dead (e.g.,
after lethal irradiation or chemotherapy aimed at nucleic acids) may retain
membrane integrity for days.

Also, one can make cells transiently permeable to trypan blue or propidium
with lysolecithin, heat or cold shock, electroporation, etc.

For further information and references, see pp. 247 and 307 of the 3rd
Edition of Practical Flow Cytometry.

-Howard