Resolution with Cy5 and PE-Cy5

From: Robinson, Howard (Howard.Robinson@joslin.harvard.edu)
Date: Mon Mar 27 2000 - 17:53:48 EST


On our MoFlo, the first pathway (488nm) has 5 PMT holders for SSC and 4 FL
channels.  The third pathway (633nm) has 2 PMT holders.  In trying to ensure
I am starting with a properly aligned machine, I use spherotech 5 peak
beads.  When I use a 670/40 or 670/30 filter to align the system to look at
Cy5 or PE-Cy5, I see mixed results.

1.) With the 670/40 or 670/30 in the last position of either pathway, a
straight path, I am having great difficulty seeing more than 2 clear peaks.
The 3 lowest peaks are not resolved at all.

2.) When I insert a 710DRLP and put the 670/40 or 670/30 in position to
receive reflected light from the dichroic I get beautiful resolution of all
peaks through the 670/40 or 670/30.  The light passed on by the dichroic is
then filtered through a 740LP and sent to the last detector in either
pathway, again this resolution is poor (I get the 2 higher peaks broad and
the lowest 3 peaks clumped together).
I have adjust the PMTs voltage and physical position in the holders,
switched the PMTs, adjusted the lasers, the dichroic, and I'm at a loss for
why this is happening.

The light reflected by the dichroic and sent through the 670/40 or 670/30
gives a
better resolved signal than the direct light passed through a 670/40 or
670/30.  Is
this normal?  Am I doing something obviously wrong?

How can I ensure the resolution for the signal passed through the 740LP is
optimal?


Thank for any assistance you can offer and thanks for your help to this
point.

Howard Robinson



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:39 EST