RE: Basic set-up for E. coli with Coulter EPICS Elite

From: gerhard nebe-von-caron (Gerhard.Nebe-von-Caron@Unilever.com)
Date: Mon Mar 20 2000 - 18:19:41 EST


Regards
Gerhard

If you attend the ISAC meeting in Montpellier you might want to attend the
tutorial on microbiology.  Below a rough guide on how to set up any cytometer
for the measurement of bacteria.

First of all buy some approx. 500nm reference beads (for example yellow-green
from Polysciences calibration grade). Trigger instrument on green fluorescence
and adjust the voltages to have the beads in the top right quadrant of a side
scatter versus green fluorescence display and optimise calibration (if
applicable). There should be a lot of nothing between the beads and the origin
of the plot.

Then set your instrument to trigger simultaneously side scatter and determine
your noise background.  Bugs will give you usually less scatter than the beads,
so again, your threshold should be quite a bit / more than one decade below the
bead cluster. On single channel triggered instruments you just have to go for
side scatter only.  Most instruments are unable to resolve bugs by their normal
forward scatter signal. If you can not separate your bugs from the noise you
have either got a dirty flow cell of lots of other bits in the sample or more
likely sheath line (contact me for advice on cleaning) . This is how you
usually find out about the first law of microbial cytometry: "sterile is not
particle free". Best use filtered liquids in disposable labware. We actually
put some disposable Millipore filter cartridges into the sheath line next to
the flow cell. Extremely useful for sterile sorting!

You can than mix beads with bugs to see them simultaneously. Ideally you start
with heatfixed bugs stained with PI as you do not have to worry about dye
extrusion pumps. However, you can also use green fluorescent bacteria from the
Phagotest (Becton Dickinson / Orpegen) or make your own green bugs using
carboxy-fluorescein-diacetate-succinimidylester.  This combination should also
allow you to set your red vs green compensation for this fitc/pi.

Supravital stains are tricky as most bugs pump like mad. The highly permeable
dyes sold by molecular probes work, but only if added fresh at the suggested
concentration. Otherwise try to start with the neat cultures,
stationary/starved e.coli (washed 4-5 times in PBS) to minimise the effect of
extrusion pumps.

E.coli from culture give quite some scatter signal. Listeria is more difficult.

To classify the functional state of your bugs see
Nebe-von Caron, G., Stephens, P., and Badley, R.A. Assessment Of Bacterial
Viability Status By Flow Cytometry And Single Cell Sorting . J.Appl.Microbiol.
84:988-998, 1998. and the revised table added below.

FUNCTIONAL	- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - INTACT
CELLS - - - - - - - - - - - - - -	- - - - - - - - - - - -		PERMEABILISED
CELL	- - - - - METABOLICALLY ACTIVE CELLS - - - - - -			(DEAD)
STATUS	REPRODUCTIVE GROWING CELLS				CELLS

TEST CRITERIA	CELL DIVISION	METABOLIC ACTIVITY	MEM. INTEGRITY		MEMBRANE
PERMEABILITY

DETECTION
METHOD	CELL COUNTING
	FIXED VOLUME COUNT
	TIME INTEGRATION
	RATIOMETRIC COUNTING
	CELL TRACKING
	COVALENT LABELLING (INTRACELLULAR OR MEMBRANE)

		ENERGY DEPENDENT
		BIOSYNTHESIS
		(DNA / PROTEIN SYNTHESIS,
		CELL ELONGATION UNDER ANTIBIOTIC
		PRESSURE; Direct Viable Count, DVC)
		PUMP ACTIVITY
		(pH CONTROL, DYE EFFLUX)
		MEMBRANE POTENTIAL
		(ACCUMULATION OF CATIONIC
		DYES, DISSIPATION OF ANIONIC DYES)
		ENERGY INDEPENDENT
		ENZYME ACTIVITY
		SUBSTRATE CONVERSION BY ESTERASE
		or DEHYDROGENASE

				SELECTIVE MEMBRANE PERMEABILITY
				EXCLUSION OF MEMBRANE IMPERMEABLE
				DNA STAINS (PI) BUT UPTAKE OF SUPRAVITAL
				DNA STAINS  (HOECHST 33342, DAPI,
				SYTO 9/13  ToPro-3, EB)

						INDISCRIMINATE UPTAKE OF
						CYTOPLASMIC STAINS
		- - - - - - - - - - - - - - - - - CELLS WITH NUCLEIC ACID DAMAGE
(non-growing)- - -  - - - - - - - - - - -- - - - - - -

Viability measurements in their functional context:
Reproductive growth as the most stringent proof for viability requires
metabolic activity and in turn membrane integrity.  In a lot of cases this
function can not be measured due to irreversible DNA damage, fastidious growth
conditions, lack of symbiotic partners or extremely slow growth.  Detection of
metabolic activity is less stringent but suggests the absence of cell death.
Whilst it does not warrant reproductive growth, this function might be
sufficient to generate unwanted effects such as food spoilage or accumulation
of toxins or transfer of genes.  In cases of injury, dormancy or extreme
starvation metabolic functions might be below detection limit.	Membrane
integrity demonstrates the protection of cell constituents and its potential to
generate gradients thus it’s capability of metabolic activity / repair and
potentially reproductive growth unless the DNA is damaged beyond repair.  It
also separates bacteria from other organic matter and debris.  Cells without an
intact membrane can not maintain any electrochemical gradient and can be
classified as dead cells. As their structures are freely exposed to the
environment they will eventually decompose.

-----Original Message-----
From:	Craig Shadbolt [SMTP:craig_ts@postoffice.utas.edu.au]
Sent:	Saturday, March 18, 2000 4:15 AM
To:	Cytometry Mailing List
Subject:	Basic setup for E. coli with Coulter EPICS Elite


Dear Flow-Experts


Being the rank flow cytometry novice that I am, I thought it best to
ask others for some advice. I am very interested in using flow
cytometry for my work involving bacterial heterogeneity during food
processing of goods such as femented meats. Unfortunately there is
little expertise here in dealing with bacteria in flow cytometers. The
machine we have (Coulter EPICS Elite) is used only for clinical
studies.


I am hoping to change that, but I am currently having trouble breaking
my cells out of the background clutter. I believe this may be due to
the stain I was employing, propidium iodide (PI). The cell cultures I
was using were a salt-stressed and a starved population of <italic>E.
coli</italic>. I suspect that most of them were still reasonably
healthy, so they would not have taken up the PI. Can anybody suggest a
simple protocol using some common stains which would enable me to
visualise my cultures in terms of viable/non-viable-injured/dead cells
and get me started on this topic?


Regards

Craig Shadbolt


++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++++


Craig Shadbolt

School of Agricultural Science and Tasmanian Institute of Agricultural
Research

University of Tasmania

GPO Box 252-54

Hobart, 7001

Australia


Phone	Intl. +61 (03) 6226 2735

Fax	Intl. +61 (03) 6226 2642

email	craig_ts@utas.edu.au



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:38 EST