Apart from probably non-specific binding via some of the fluorochromes there is
always the possibility of cell cell aggregation, especially in non-healthy
people. The numbers of cells that aggregated seem to be higher in preparation
methods that involve pelleting steps.  I had a poster about this on Toronto a
few years ago. As these days the numbers of plots displayed is not limiting any
more I would always suggest to look at log side scatter versus each of the
fluorescence channels used as a basic form of display for general sample
assessment. Log side scatter allows to visualise lymphocytes, monocytes
neutrophils and eosinophils (and sometimes a few more things) in particular
after FACSlyse . Aggregates between monocytes and lymphocytes are easily
identified and in most cases there will also be some granulocytes showing for
example CD19 as well.  In the case I reported it was interesting that the
coaggregation was abolished using anti CD45. Thus the classical confirmation of
the lymphocyte scatter gate became completely invalid.

Regards
Gerhard

-----Original Message-----
From:	Annuska Glas [SMTP:aglas@vmresearch.org]
Sent:	Tuesday, March 07, 2000 11:11 PM
To:	Cytometry Mailing List
Subject:	CD14 and CD19 positive cells


Hello,

I was wondering if anyone has seen a population positive for both CD14 and
CD19. I am staining peripheral blood lymphocytes and am looking at the B cell
population. I originally just used CD19 to get the B cells. However recently I
began using CD3, CD14 to exclude any T cells and monocytes. I will use CD56 as
well to exclude NK cells. However, I did see some cells that were positive for
both CD19 and CD14. Are these cells monocytes or are they B cells. I do want to
look at all B cells, my question is how can I find out if these cells are truly
B cells or if they are monocytes?

Thanks,
Annuska Glas