RE: Re: Hoechst and microbiology

From: Joerg Ueckert (Joerg.Ueckert@Unilever.com)
Date: Mon Mar 06 2000 - 08:36:35 EST


Sue,

we have similar experiences with fluorescence variability among bacterial
species as described by Howard, also, with repect to efflux pumps.
In addition, a remark on the UV- laser output: In the past we used a FACS II
(YES, some of them are still alive) with a 200 mW water-cooled laser for
Hoechst 33342 fluorescence in live bacteria; we got clear fluorescence
separation of unstained / stained populations. In contrast, the 20 mW 323nm
air-cooled laser of our Elite now gives big overlaps of pos./negatives. We came
to the point that we somewhat question the use of Hoechst 33342 as live stain
for bacteria, especially with low-output UV-lasers or in combination with (0.1
µm-filtered) minimal media; the number of events accounting for background
fluorescence is always hard to assess.

Joerg Ueckert
Unilever Research, NL


-----Original Message-----
From:	Howard Shapiro [SMTP:hms@shapirolab.com]
Sent:	Friday, February 18, 2000 1:23 AM
To:	Cytometry Mailing List
Subject:	Re: Hoechst and microbiology


Sue DeMaggio asks:


>Has anyone used Hoechst to stain bacteria?  or have a procedure for it?  I
>have a client asking.

Live or fixed?	Hoechst will get into some live bacteria; so will DAPI;
both will stain a broader range of bacteria after ethanol fixation.  Many
years ago, the Livermore group used staining with Hoechst 33258 and
chromomycin A3 to discriminate bacterial species on the basis of the
A-T/G-C ratio in DNA (Van Dilla MA, Langlois RG, Pinkel D et al: Bacterial
characterization by flow cytometry.  Science 220:620, 1983).  However, a
recent paper (Walberg M, Gaustad P, Steen HB: Uptake kinetics of nucleic
acid targeting dyes in S. aureus, E. faecalis, and B. Cereus: a flow
cytometric study. J Microbiol Methods 1999; 35:167-176) indicates that
staining of bacteria by Hoechst dyes and most other nucleic acid stains can
vary considerably from species to species, and may be affected by the
presence or absence of efflux pumps (in live cells), as well as by the
fixative or permeabilizing agent used.	With a consistent preparative
procedure, one ought to be able to compare DNA contents of different cells
from a single species; comparing species could be trickier.  It is becoming
increasingly evident that, in terms of their behavior when exposed to the
fluorescent dyes we know and love, bacteria aren't just little eukaryotes.

-Howard