you need to remember that the CD34 cells are considered stem cells
and from
the data available for Rhodamine 123 and Hoechst staining of these cells,
they exhibit a highly active MDR system.  It is possible that the
cells are
pumping the indo out as fast as it is going in.

Charles A. Kuszynski, Ph.D.