A little more on CFSE

From: Lisa Gale (lisa.gale@adelaide.edu.au)
Date: Wed Feb 23 2000 - 18:33:13 EST


Hi, I thought that since there was so much interest in CFSE I would
share my recent experiences. I hope it will be of some use.

I have been labelling human PBMC with CFSE according to Current
Protocols methodology. After labelling, my cells were more than 95%
viable by trypan blue exclusion... but would not proliferate in culture.
In the end, by titration, I found that reducing the final concentration
of CFSE to 2 micromolar, (instead of the suggested 10 micromolar) gave
good staining while still allowing good proliferation.

So the moral of the story seems to be that CSFE can do funny things to
cells, and although they are viable they may not be fully functional!


--
Lisa M Gale