I would like to add a few comments to the recent discussion on CSFE 
labeling lymphocytes for proliferation analysis.

1.	We adopted the protocol for bovine PBMC and have done quite an extensive 
analysis of differential responses of lymphocyte subpopulations in response 
to stimulation by a variety of mitogens ( ConA, PMA/Ionomycin, Anti-CD3 
antibody, Anti-delta antibody and r-IL 2).  The CSFE labeling system worked 
very well for our analysis and gave us valuable information about which 
sub-population are responding primarily, which ones respond as by-standers, 
what proportion (%) of each sub-population is dividing (at each cell 
division level) and also allowed us a time course analysis of 
response.  All these differences were not detectable by the conventional 3H 
tritiated thymidine incoporation assay.

	I have attached a MS powerpoint file of a portion of one such experiment 
for a preview of those who are interested.

2.	We prepared CSFE as a 20mM solution in DMSO and froze ~10-20 ul aliquots 
at -20*C. We were using it for more than an year now with out any 
detectable change in labeling activity.  We use a dye dispersing agent 
called "Pluronic F127" (Molecular probes) supplied as 20% solution in DMSO, 
to help loading of CFSE very efficiently.  We thaw out a 20 mM CSFE vial 
and mix with Pluronic F127 at 1:1 ratio, and then dilute the mixture in 
Hanks Balanced Salt Solution to 1.5-3 micro molar (uM) concentration and 
filter sterilize it before labeling.  Use of Pluronic F127 allowed us to 
use the CSFE at very low concentrations ( bovine PBMC were very sensitive 
to CSFE and started dying at or above 5 uM concentrations) and gave an 
excellent loading of the CFSE in to the cytoplasm.

3.	We noticed that for first 12 hours in culture, the cells were very 
bright for flowcytometry analysis. After initial quenching of some dye out 
from cells, after 12 hours onwards we were able to compensate the 
green(FL1) and red(FL2) perfectly for our two color analysis.  We use to 
set FL2-FL1 compensations between 90 and 100% in order to set the 
compensation properly.  We also had to individually titrate  all our FL2 
(PE) antibodies initially for surface staining for phenotype analysis 
to  have them in a viewable level.

Hope this information will be useful for those who are using CSFE analysis.