A means by which dead cells can be discriminated from live cells, when both are labeled with CFSE, is to use 7-AAD. Only dead cells are permeable to 7-AAD and its emission peak is far enough removed from the broad range of CFSE emission to be easily compensated. Kevin G. Waddick, Ph.D. Parker Hughes Institute tessa kerre wrote: > Dear flowers, > > In addition to Dr. Waddick's questions concerning CFSE, we too have some > problems with it. > > I have been using CFSE to track human cord blood HSC in vivo and check > for cell divisions. > > - I cannot use frozen material, because then I get clump forming, which > results in loss of 75% of my cells. Is this normal? > - It is extremely difficult to compensate the overflow from CFSE towards > the FL-3 and the FL-2 channel, making it difficult top combine CFSE with > PE or Cy labeled antibodies. Gating out of dead cells using propidium > iodide also becomes very difficult. Do others also have the same > problems, or do you have the magical settings? > - How long can you keep CFSE in good shape frozen in DMSO (-20°C)? > > Kind regards, > > Tessa