Re: your mail

From: Joseph Trotter (trotter@scripps.edu)
Date: Wed Jan 19 2000 - 15:33:26 EST


Diane,

	One normally usually uses FL2 to measure the PI signal (the
PE channel - 585/42BP), but many yeast folks use FL3. This is because
there is a lot less PI signal in FL2 than in FL3 on *most* cytometers,
allowing for more high voltage and better linearity for the DNA
measurement. With yeast, the smaller DNA signal makes it worth
your while to check both FL2 and FL3 in your system to determine
the best channel to use. At Scripps, most yeast PI signal is 
measured in FL3 on the FACScans (650nm LP), and in FL2 on a
FACS Calibur (FL3 is a 670nm LP on the new systems).

	Since you're measuring nuclear fluorescence, and the
geometry of the cells may be unusual depending on the yeast
you are assaying, simple pulse height may produce better
results than pulse area. This is almost never true for
mammalian cells, but often true for yeast. Use an amplifier
gain that is low and as much high voltage as reasonable for
the best signals. In addition, match the sample and sheath buffers.
Some yeast protocols use citrate buffers that have a
different refractive index than PBS, producing optical
artifacts you don't need. Some of our yeast folks bring
their own sheath buffer, and some use PBS sample buffer to
accomplish this. 

	Good luck,

		Joe

Joseph Trotter, Director
Flow Cytometry Core Facility
The Scripps Research Institute
10550 North Torrey Pines Road
La Jolla, CA 92037

On Tue, 18 Jan 2000, Diane M Sharp wrote:

> Hello all,
> 
> I have recently been asked to run cell cycle on yeast.  I have run endless
> samples for cell cycle on tumor lines, but never on yeast.  I will be
> using propidium iodide.  Is this a difficult assay?  I would appreciate
> any comments from folks who have already run this type of sample.
> Thanks in advance for your help!
> 
> Diane Sharp 
> DuPont Pharmaceuticals
> Glenolden, PA
> 



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:55:27 EST