Re: Expo software

From: Mario Roederer (Roederer@drmr.com)
Date: Mon Jan 10 2000 - 16:44:26 EST


Mark,

Autofluorescence compensation was published back in the mid-1980's, 
doing it both by software (Roederer & Murphy, Cytometry) and by 
hardware (Alberti, Parks & Herzenberg, Cytometry).  Basically, you 
just treat one of your fluorescence channels (for which there is no 
stain, for example, FL2 in the absence of a PE stain) as the channel 
that measures autofluorescence.  Then you just do standard 
compensation based on removing the spillover between this channel and 
the others.  There is no need to collect data at different voltages 
(but if you do, make sure to collect the "compensation controls" 
(including the unstained autofluorescence control) at each different 
voltage and adjust comp settings for each voltages separately).

For Macintosh, FlowJo can perform this compensation on your data, 
assuming you have collected all of the appropriate compensation 
controls already.  For PC, a variety of programs will let you do this.

I don't think you should use a generic "algebraic" approach, because 
what you are doing is compensation, NOT background subtraction. 
Therefore, you need to use the linear algebra mathematics that are 
built into the compensation algorithms of these programs.

mr

>I have a separate question: "After-the-fact compensation"
>
>     We have numerous FCS files (generated on a Coulter Elite using Expo
>software but typically analyzed using CellQuest).  We would like to try
>different kinds of compensation, after the experiment is over.  The object
>is primarily to remove autofluorescence in order to improve the signal to
>noise ratio of our samples.  In other words we would like to run "induced"
>and uninduced" samples at various voltage settings for the different PMTs
>and, at a later point, try correcting for autofluorescence by subtracting a
>fraction of one channel from another channel.
>     Ideally this would be a program which inputs one or more FCS files
>together with an algebraic expression representing the operation to be done
>on the files and outputs FCS files containing the altered parameter (or
>maybe a new parameter representing the transformed data).  The only trick
>might be that log data would have to be converted to a linear signal before
>subtraction and then converted back to log data.
>
>Mark Poritz, Ph.D.