Re: CD34 enumeration - Reposted

From: Mike Keeney (Mike.Keeney@LHSC.ON.CA)
Date: Tue Jan 04 2000 - 16:23:40 EST


[NOTE: I'm reposting this because it apparently got scrambled somewhere ]
[in our system.  Steve                                                  ]

Dear Andrea,

In response to your questions:
1. which procedure is recommended these days?
There are several methods of enumerating CD34 commercially available and 
the one you choose depends on your own specific requirements. The IMAGN 
2000 from BD is a microvolume fluorimeter (non flow based) method which is
perhaps the simplest to use. Recent report show good C.V.s on blood and 
apheresis. Limitations are high C.V.s in low CD34 count range (<10/ul)due 
to the limited volume analysed and as it is a single color system the 
effect of low viabilty is undocumented.

The ProCount from BD is a 3 color single platform method utilizing a vital 
dye on FL1, CD34PE and CD45PerCp. The addition of TruCOUNT fluorospheres 
allows absolute CD34 numbers to be derived directly from the flow 
cytometer. Published studies also show good C.V.s for this method and the 
availabilty of an automated algorithm may improve reproducibilty between 
sites. Limitations are the inabilty to measure viabilty and because all 3 
PMT's are used in this method it is not possible to do CD34 subset 
analysis on a 3 color laser. Apheresis packs with many platelet clumps may 
cause a problem for the automated software, and a manual gating strategy 
using either ProCOUNT or ISHAGE is available.

The StemKIT from Immunotech/Beckman-Coulter is a 2 color CD45FITC CD34PE 
method that uses FlowCOUNT fluorospheres to allow single platform CD34 
absolute counts to be determined. Currently gating is performed manually 
using the ISHAGE guidlines. An advantage of this method is the ability to 
include 7-AAD in FL3 (or FL4) to assess viabilty. Subsets of CD34 can also 
be determined by the addition of a 3rd antibody.

Recent reports from the European Working Group on Clinical Cell Analysis 
show very good inter and intra laboratory C.V.'s with this method.

2. which antibody is one supposed to use?  Are the ISEHAGE guidelines 
still recommending CD34-PE?

For 2 color analysis CD34PE is the antibody of choice because of its 
excellent signal/noise ratio. As long as class III antibodies of CD34 are 
used FITC conjugates also work well. Other fluorochromes can be used but 
each fluorochrome/antibody combination should be evaluated before 
implementing in a clinical lab.

3. are most people just doing the single parameter CD34 percentage or is 
the lineage-restricted vs non-restricted %CD34 analysis becoming more 
important.

>From the research viewpoint there is tremendous interest in subsets of 
CD34. However most clinical data published to date confirms that the 
single most important factor predicting both short and long term engraftment
is related to the total number of CD34+/kg body weight infused into the 
patient.

Back to your initial problem. Do you wash your samples after lysing? Most 
experts in the field now recommend no wash techniques.

Are the absolute numbers as well as the percentages different? One or 
other method may be causing a selective loss of a specific cell type.
How does the other lab measure their CD34+ cells? It is safe to measure 
CD34 cells without lysing in apheresis packs as the number of contaminating
 rbc's is usually low. Simply resuspend the stained sample in 1ml of PBS 
before analysis. This is not recommended for whole blood or cord blood due 
to the very high number of red blood cells.

I hope this helps. Some excellent references can be found in the Journal 
of Hematotherapy, Transfusion and Clinical Cytometry. Many of the points 
you raised have also recently been covered in Current Protocols in 
Cytometry by Jan Gratama, Rob Sutherland and myself.
Feel free to contact me directly if you wish followup on any of the 
points.

Good Luck!!

Mike Keeney



<<< "Andrea Illingworth" <dcdsflow@mint.net>  1/ 3  1:01p >>>
Dear flow group,
We are currently in the process of evaluating a reliable procedure for 
CD34 enumeration.  We have been sharing samples and comparing results with 
another institution and have been surprised to find out in the process 
that lysed specimens yield consistently higher percentages than unlysed 
specimens.  We have been using the Coulter Stemkit which uses CD34-PE/CD45-
FITC and their designated lysing solution.  The other lab is using a 
no-lysis procedure and a CD45-FITC/CD34-PerCP combination.

I am sure I am not the first one to observe this finding and my questions 
are :
1. which procedure is recommended these days?
2. which antibody is one supposed to use?  Are the ISEHAGE guidelines 
still recommending CD34-PE?
3. are most people just doing the single parameter CD34 percentage or is 
the lineage-restricted vs non-restricted %CD34 analysis becoming more 
important.

Thanks for your help and have a happy and safe change into the new 
millennium!
Andrea Illingworth, MS, H(ASCP)
Dahl-Chase Diagnostic Services
Flow Cytometry
333 State Street
Bangor, Maine 04401