Re: beta-gal

From: Mario Roederer (Roederer@drmr.com)
Date: Mon Oct 25 1999 - 18:14:39 EST


Jackie,

while no longer in vogue (due to the popularity/ease of GFP), 
beta-gal remains an excellent choice for reporter gene detection.  It 
is considerably more sensitive than GFP; we were able to detect as 
few as 5 molecules of beta-gal per cell: probably a single mRNA 
molecule's worth!  However, the beta-gal assay requires just a little 
bit more work than GFP.  A full description of the methodology can be 
found in a few references.  Also see the chapter in Current Protocols 
for Cytometry, which updates a review of these articles plus a 
variety of articles that have used the method in the past 10 years.

1. Nolan, G. P., et al. (1988).  Fluorescence-activated cell analysis 
and sorting of viable mammalian cells based on beta-D-galactosidase 
activity after transduction of Escherichia coli lacZ.  Proc Natl Acad 
Sci U S A 85:2603.
2. Fiering, S., et al. (1991).  Improved FACS-Gal: flow cytometric 
analysis and sorting of viable eukaryotic cells expressing reporter 
gene constructs.  Cytometry 12:291.
3. Roederer, M., et al. (1991).  FACS-Gal:  Flow cytometric analysis 
and sorting of cells expressing reporter gene constructs.  Methods: A 
Companion to Methods in Enzymology 2:248.