FRET

From: Kathleen Schell (kschell@facstaff.wisc.edu)
Date: Tue Aug 03 1999 - 11:46:35 EST


Hi, flowers,
     We are beginning to do some experimentation with FRET and I have some
instrumentation questions.  First, the experiment:  E. coli has been
transfected with both GFP + proteinA and also with BFP+proteinB.  Both
proteins are membrane-bound and don't come together unless a stimulus is
applied.
    We have been using a 488 (argon) in the primary postion and a 405
(krypton) in the 2nd position to detect these two fluorophores on a
FACStarPlus.  We have been looking at the GFP (530/30) alone off the 1st
laser and the BFP (460/20) off the second.  In addition, we have a detector
that looks at  any GFP generated by FRET off the second (also a
530/30--with a 505 beam splitter).  Because the expression is low we have
been collecting signals with linear amplification--requiring increased gain
settings.
Question 1:  Do we have the right instrument configuration to collect the
respective signals?  The blue we see (when we can see it, is very dim).,
but the green is significant.  (Stupid question --will exciting the GFP
with 488 first, render it unexcitable by the BFP if the beams are spatially
separated?)
Question 2:  We are also collecting a ratio (GFP-second beam/BFP-second
beam) but have been having some difficulty because the blue signal has been
so dim.  Is this an appropriate measurement?
Question 3:  If the transfection efficiency of the BFP-proteinB combination
is bad, can I assume we are SOL?  Should the signal intensity of both
fluorophores be somewhat equal to get good measurements?
     Any suggestions would be greatly appreciated.
     Thanks in advance.
Kathy



--------------------------------------------------------------