Dear Claudio,

In our institute the departement of Clinical Viro-Immunology studied the
phenotype of effector T cells, using among others Perforine and Granzyme
antibodies.
Publication; Hamann et all, J.Ex.Med 1997, 186, 1407-1418.

The monoclonal antibody to Perforine was obtained from of Holzel
Diagnostika, Koln, Germany and the Granzyme-B antibody is produced in our
own institute by CLB-Reagents (commercially available).

CLB protocol for membrane and intracellular FACS staining :
(By Paul Baars)

Reagents:
PBS
PBS 0.5% BSA
PBS 0.1% saponin 0.5% BSA
Human Pooled Serum (HPS)
4% Paraformaldehyde (PFA) in PBS


Procedure (the whole procedure is performed on ice)
1.	Wash the cells in a 15 ml tube in PBS 0.5% BSA
2.	Suspend the cells in PBS 0.5% BSA (4x106/ml) and add direct
conjugated Mab's for the membrane staining
3.	Incubate for 30 min
4.	Wash 1X with PBS 0.5% BSA
5.	Wash 1X with PBS
6.	Add 1.5 ml 4% PFA in PBS and incubate for 5 min (stopwatch!!)
7.	Wash 1X with PBS
8.	Wash 1X with PBS 0.1% saponin 0.5% BSA
9.	Suspend the cells in PBS 0.1% saponin 0.5% BSA + 10% HPS 
10.	Incubate for 20 min
11.	Wash 1X with PBS 0.1% saponin 0.5% BSA
12.	(From now on this buffer is used untill the end of the procedure)
13.	Suspend the cells in a x 50ml (a= number of different intracellular
stainings) and pipette the cells in a 96-well dish
14.	Add Mab's to the intracellular antigen and control Mab's (diluted in
saponin buffer)
15.	Incubate for 30 min
16.	Wash 3X
17.	Measure the cells on the FACS



We at CLB which you success with the protocol, If you want example
histograms please let me know so we can send them by separate E-mail.

Best regards,

John Voorn

CLB Reagents	tel: 	+31-20-5123246	
Plesmanlaan 125	fax: 	+311-20-5123570	
1066CX Amsterdam	E-mail:	J_Voorn@CLB.nl <mailto:J_Voorn@CLB.nl> 	
the Netherlands	Site:	WWW.CLB.nl <http://WWW.CLB.nl> 	



	-----Original Message-----
	From:	claudio.vallan@dkf7.unibe.ch
[SMTP:claudio.vallan@dkf7.unibe.ch]
	Sent:	maandag 19 april 1999 10:53
	To:	cytometry@flowcyt.cyto.purdue.edu
	Subject:	intracellular perforin staining




	Received: from ns-clb.clb.nl by ln210-2.clb.nl (Lotus SMTP MTA v1.05
(274.9
	 11-27-1996)) with SMTP id C1256758.005B4F60; Mon, 19 Apr 1999
17:39:22
	+0200
	Received: from flowcyt.cyto.purdue.edu by ns-clb.clb.nl (AIX 4.1/UCB
	5.64/4.03)
	          id AA24492; Mon, 19 Apr 1999 18:47:07 +0200
	Received: (from root@localhost)
	 by flowcyt.cyto.purdue.edu (8.9.3/8.9.3) id JAA07571
	 for cyto-sendout; Mon, 19 Apr 1999 09:51:01 -0500
	Received: (from daemon@localhost)
	 by flowcyt.cyto.purdue.edu (8.9.3/8.9.3) id DAA06133
	 for kelley; Mon, 19 Apr 1999 03:47:12 -0500
	Received: from arwen.unibe.ch (arwen.unibe.ch [130.92.9.52])
	 by flowcyt.cyto.purdue.edu (8.9.3/8.9.3) with SMTP id DAA06129
	 for <cytometry@flowcyt.cyto.purdue.edu>; Mon, 19 Apr 1999 03:47:10
-0500
	Received: from ubecx01 (actually ubecx01.unibe.ch) by arwen with
smtpL;
	          Mon, 19 Apr 1999 10:49:15 +0200
	Received: from [130.92.148.110] (pathomac110.unibe.ch)
	          by ubecx01.unibe.ch (PMDF V5.1-12 #21734) with ESMTP
	          id <0FAF009REHZX0H@ubecx01.unibe.ch>
	          for cytometry@flowcyt.cyto.purdue.edu;
	          Mon, 19 Apr 1999 10:52:46 +0200 (MET DST)
	Date: Mon, 19 Apr 1999 10:52:46 +0100
	From: Claudio Vallan <claudio.vallan@dkf7.unibe.ch>
	Subject: intracellular perforin staining
	X-Sender: vallan@ubecx01.unibe.ch
	To: Cytometry Mailing List <cytometry@flowcyt.cyto.purdue.edu>
	Message-Id: <l03110700b340ac7d1aed@[130.92.148.110]>
	Mime-Version: 1.0
	Content-Type: text/plain; charset="us-ascii"




	Does anybody have a protocol for intacellular staining of human
perforin
	that works?  I used the Ab from Pharmingen on activated
intraepithelial
	lymphocytes of the small intestine and two different methods for
	permeabilization: PFA + saponin and OrthoPermeaFix. As positive
control I
	used CTL clones. The Ab gave high background but no specific
staining.
	Thanks for your help!

	



This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:53:23 EST