Re: Sorting C. elegans embryos

From: Saverio Alberti (alberti@cmns.mnegri.it)
Date: Tue Mar 09 1999 - 12:43:21 EST


given the size of the embryo vs the diameter of the orifice, you
probably are severely disturbing the drop formation process. An old (sorry
Dick) article by Dick Stovell, Len and others has demonstrated that even
when sorting normal cells, i. e. at much more favourable ratios of cell
diameter vs orifice diameter.

Saverio Alberti
Head, Unit of Experimental Oncology
Department of Cell Biology and Oncology
Consorzio Mario Negri Sud
66030 Santa Maria Imbaro (Chieti), Italy
Phone: (39) 872-570.293
FAX: (39) 872-578.240
E-mail: alberti@cmns.mnegri.it


On Mon, 8 Mar 1999, Wayne Green wrote:

> 
> 	An investigator here wants to sort C. elegans embryos expressing
> GFP.  The embryos are anywhere between the 28 cell to 600 cell stage with
> most closer to the smaller size.  Each embryo is roughly 50x30 microns,
> ellipsoid in shape with a fairly rigid shell encasing the embryo.  We are
> told that the embryos are relatively resistant to physical trauma.
> 	We have run these on a Vantage with a 100 micron tip (no Macro Sort)
> and can detect the GFP expressing embryos.  Our problem arises with the
> sorting; if we sort 40,000 GFP positive embryos, one only finds about 100 in
> the collection tube.  The good news is that those 100 are healthy in that
> they grow normally following the sorting process.  	The question is,
> what might cause this abysmal sorting efficiency?  When we sort "normal"
> cells like murine stem cells, lymphocytes, etc. we get very good recovery
> and purity.  If there is anyone out there with experience sorting something
> like this I would greatly appreciate your suggestions.
> 
> Thanks
> 
> Wayne F. Green, Ph.D.
> HCI Flow Cytometry Core Lab
> wayne.green@hci.utah.edu
> 
> 