One of our users has sorted based on DNA content. They use 10ug/ml Hoechst 342 and stain for one hour at 37C. With adherent cultures, they have the cells in log phase (50-60% confluent). With suspension cultures they stain at no more than 3million/ml, but the cells are grown at 500,000 to 800,000/ml then concentrated for the staining. After the cells are harvested they are resuspended in media with 10ug/ml H342 at 10 to 20 million/ml. The sample block is kept cold during the sort and the cells are sorted into tubes that are also on ice to block progression thru the cell cycle. Our user isolates the DNA from the cells after the sort so I can't respond to the viablity of the cells at that point. It is important that your UV power is low (50-100mW) so you don't fry the cells. You should check that the time of staining and cell concentration during the staining are optimal for your cell line and you can check at that time if your cells are still viable. Good Luck. Lucy Brown Beckman Research Institute/City of Hope 626-359-8111 X3306 lbrown@coh.org