Arnold, I'm not sure I understand what you really want to do, here is what is possible: The Cellquest listmode files are "pretty much" FCS 2.0 standard, so you can: 1. Copy them to a PC formatted diskette, Zip, MO or other cartridge on the Mac using PC Exchange, and transfer it to the PC 2. Use any network to transfer, but make sure you use binary transfer (or binhex, zip or uuencode the files if you need to do ASCII transfer. 3. Once transfered you can read these files with WinMDI, however you cannot read them with PClysis, because this only reads FCS 1.0 version files. I am not aware of any utility that can convert FCS 2.0 files back into FCS 1.0 files readable by PC Lysis (but why working with PCLysis when WinMDI is so much better ?). If this doesn't answer your questions please let me know, Cheers, Matthias _____________________________________________________________________________ Matthias Haury Flowcytometry Dept Immunology Institut Pasteur mhaury@pasteur.fr Tel: 33 (01) 40 61 31 29 Fax: 33 (01) 45 68 86 39 _____________________________________________________________________________ At 7:21 -0700 8/10/90, arubin@csnet.net wrote: >Does anyone know of software that can translate list mode data >from a Mac computer on a FACSCAN to a PC format similar to what >HPREADER can do with PCLYSYS? Thanks for the help. >Arnold D. Rubin,MD,St. Joseph's Hospital and Medical Center, >Paterson,NJ
This archive was generated by hypermail 2b29 : Wed Apr 03 2002 - 11:51:41 EST