Re: Counting Microspheres by Flow

From: Gerhard Nebe-von-Caron (Gerhard.Nebe-von-Caron@unilever.com)
Date: Wed Jan 14 1998 - 05:01:24 EST


          We regularly count 320nm latex beads and bacteria on the 
          EPICS XL.  We use a ratiometric method with yellow/green 
          reference beads (500nm).  Final concentrations can be 
          calculated by the volume mixing ratio and the concentration 
          of the reference beads.  The advantage of the method is that 
          it is not critical for system dead-times and variation in 
          flow rates. It also serves as on-line QC. Add detergent 
          (0.05% Tween 20) to avoid stickiness to the tubing.
          
          Regards
          Gerhard.nebe-von-caron@unilever.com


______________________________ Reply Separator _________________________________
Subject: Counting Microspheres by Flow
Author:  Padma_K_Narayanan@sbphrd.com at INTERNET
Date:    13/01/98 21:00


Hello flowers:
Is any body out there in flow land who routinely counts fluorescent
microsphere (beads) using a Cytoron. Would like to get some input about
arriving at final numbers.
Thanks
Padma

SmithKline Beecham