We have also had the experience of decreased PAC-1 binding in response to increasing concentrations of platelet agonists.These cells were still able to increase binding of CD62 and CD63 antibodies at the same time. According to Shattil (Blood, Vol 70, No 1 (July), 1987:pp 307-315) PAC-1 and fibrinogen inhibit each others binding in a competitive manner and bind to the same RDG binding site. The PAC1 antibody has a much higher affinity (Kd=5nmol/l) for the site than does fibrinogen (Kd=250nmol/l) and wins out. Shattil also mentions that he also sees a time dependent decrease in PAC-1 binding when the antibody was added after stimulation of the cells.He surmises that this occurs because fibrinogen becomes irreversibly bound to the receptor and can not be displaced by PAC-1. It is also known that bound fibrinogen, attached to its receptor, can be internalized into alpha granules over time. Proteolysis of the receptor is also a possibility. Because of these concerns, measuring fibrinogen receptor activation with PAC-1 must be interperted carefully. Alternate methods of assessing fibrinogen receptor competency are to use a LIBS antibody to a receptor which becomes expressed on IIbIIIa when it is RDG site is occupied or to use a RIBS antibody whose receptor is expressed only on fibrinogen that has bound to its receptor. Good luck and have fun, Ann Byrne abyrne@compucyte.com